Afterpulsing #51
-
|
I can't seem to get rid of the afterpulsing signal using SyncDeadTime under PH330_Edge.ini. Any idea on which set of number to try or which other settings I need to set? Current settings: [All_Channels] |
Beta Was this translation helpful? Give feedback.
Replies: 2 comments 1 reply
-
|
Hi Seung, What type of detector you are using? Usually SPADs have afterpulsing due to their structure. Can you adjust the deadtime of the channel? |
Beta Was this translation helpful? Give feedback.
-
|
I'm doing single-molecule experiment with 10-40Mhz rep rate. The detector I'm using now is the Picoquant PMA Hybrid Series. Fluorescence lifetime is around 1 ns. |
Beta Was this translation helpful? Give feedback.
Most probably you are not seeing afterpulsing but something else. The PMA Hybrid does not show any significant afterpulsing. I would say that your problem is not Snapi related but most probably due to reflections (optical or electrical)
Please contact support@picoqaunt.com mentioning your TCSPC device serial number and provide a sketch of your experimental setup (optical and electrical configuration)